Glad to hear it worked out. Taylor changed their bottle labeling recently. Reagents with a lot number only were manufactured PRIOR to 2015 so those reagents could have been 2+ years old. The DPD reagent (R-0002) is definitely susceptible to aging where the dye molecule will slowly oxidize over time to form an imine molecule which is a clear compound. Taylor puts additives into the R-0002 to stabilize the chemistry as much as possible but it's not perfect. This is why you should always keep your reagents tightly capped and stored in a cool, dry environment.
My guess is, if you purchase the titrating reagent, you'll find the "squeeze" of the bottles to be a little different and so you'll probably see some variation in droplet count (minor) due to that. I would simply suggest you pick one bottle type you like and stick to it or else you'll run yourself in circles with testing. Developing consistency in your testing skills is the most important thing.
Thanks for bringing this up. It's a very good object lesson in testing and reagent shelf-life.
My guess is, if you purchase the titrating reagent, you'll find the "squeeze" of the bottles to be a little different and so you'll probably see some variation in droplet count (minor) due to that. I would simply suggest you pick one bottle type you like and stick to it or else you'll run yourself in circles with testing. Developing consistency in your testing skills is the most important thing.