I haven't seen too much made of this in the forums, but I decided to do a test case today. Using the speed stir I conducted the test by putting the powder in and then without any delay applying the titration drops, a little faster than 1 drop per second. I got a reading of 7 ppm (7 drops). I'm testing for bromine and using the R-0872 reagent with 12.5 ml of water, so 1 ppm per drop. I then repeated the test using a timer. I put the powder in and waited exactly 2 minutes before applying the titration drops as in the first test. The second result was 14 ppm (14 drops). This indicates that even a small delay in starting the drops could throw the results off high. I'm guessing that the ratio would be about the same with chlorine.
My previous results before I obtained the speed stir would have been consistently higher than with the speed stir because, no matter how efficient you get with swirling it takes quite a bit longer to complete the test. It seems that testing procedure is nearly as important as the reagents to get consistent results. Comparing readings from different people may not be comparing apples to apples. Maybe all this is obvious to you experienced folks.
My previous results before I obtained the speed stir would have been consistently higher than with the speed stir because, no matter how efficient you get with swirling it takes quite a bit longer to complete the test. It seems that testing procedure is nearly as important as the reagents to get consistent results. Comparing readings from different people may not be comparing apples to apples. Maybe all this is obvious to you experienced folks.