So I work in a biological research lab and have access to a laboratory grade probe-style pH meter (Mettler Toledo FE20). Out of curiosity, for the past few pH measurements that I did for my pool (chlorine) and my spa (bromine) using my Taylor K2006C, I also brought a sample in with me to measure using the lab pH meter. I would always calibrate the pH meter with fresh reagents for a curve between pH 7 and 10. Interestingly I am always getting a much lower pH measurement with the lab pH meter compared to the Taylor kit readings.
For example with todays test, using the Taylor kit, my pool measured at 7.5 and my spa measured at 7.6. However, using the lab meter, I was getting 6.95 for the pool and 7.05 for the spa.
I was just wondering why the discrepancy. I am hopeful that the Taylor measurements are correct and the kit is specifically designed for use with pool/spa water. However, the lab grade pH meter is a more sophisticated instrument that has an accuracy range of -/+ 0.01. My only thought is that the presence of chlorine or bromine obscures the lab meter results, but any other ideas?
For example with todays test, using the Taylor kit, my pool measured at 7.5 and my spa measured at 7.6. However, using the lab meter, I was getting 6.95 for the pool and 7.05 for the spa.
I was just wondering why the discrepancy. I am hopeful that the Taylor measurements are correct and the kit is specifically designed for use with pool/spa water. However, the lab grade pH meter is a more sophisticated instrument that has an accuracy range of -/+ 0.01. My only thought is that the presence of chlorine or bromine obscures the lab meter results, but any other ideas?