    This has me puzzled, and it has happened a few times. A few days ago I tested chlorine. I had been letting it stay at about 7.5 ppm for several days. After a full day of consistent rain, I tested the water expecting to see a lower result, but the test gave me a result of 9.5 ppm. Since that made no sense to me, I used water from the same sample for a second test with a result of 5.5 ppm.

    Today I tested and first got a result of 10 ppm, tried it again with a result of 8.5 ppm, and a third time with a result of 7.5 ppm, all with water from the same sample. My other test today resulted in CC 0 and pH 7.6. Any ideas what's going on?

    If it makes a difference, my full set of test results from 12 days ago was
    FC 8
    CC O
    pH 7.8
    TA 70
    CH 330
    CYA 80
    Salt 3200
    Temp 78
    CSI -0.17
    16K gallon IG exposed aggregate pool with Aquarite SWG T15 cell, Hayward C1750 filter, EcoStar SP3400VSP pump, H400FD pool/spa heater, Polaris PB4-60 booster pump, 280 pressure cleaner, Hayward Super II spa pump and Air Supply Silencer spa blower

  2. Back To Top    #2

    Join Date
    Aug 2014
    Location
    Shallotte, NC
    Posts
    693

    Re: Inconsistent FAS-DPD Test

    Same water, different results is sounding like testing error. Check out this link:

    http://www.troublefreepool.com/threa...Kit-Directions



  3. Back To Top    #3
    epro05's Avatar
    Join Date
    Jun 2014
    Location
    Keller, Texas
    Posts
    270

    Re: Inconsistent FAS-DPD Test

    Just a few thoughts.

    For consistency in drop size, make sure the reagent bottle is held perfectly vertical when you are dispensing the drops and shoot for about 1 drop per second. Make sure the hand you're using to dispense the drops is motionless, and not shaking if you are hand swirling (speed stir make this a lot easier). Make sure you are rinsing the test vial thoroughly (using pool water sample to rinse) before and after each of your tests.
    27,500 gallon in ground gunite/plaster (1995). Attached Spa with fountain, spillway, and 2 lion heads streaming into the pool. Hayward pump and AO Smith UST1152 1.5 hp motor (single speed). Hayward 60 sq ft DE Filter (Model DE-6000). Kreepy Krauly cleaner.

  4. Back To Top    #4

    Join Date
    Jul 2015
    Location
    Clearwater, FL
    Posts
    83

    Re: Inconsistent FAS-DPD Test

    Following directions to a T. That's why it's so puzzling.
    16K gallon IG exposed aggregate pool with Aquarite SWG T15 cell, Hayward C1750 filter, EcoStar SP3400VSP pump, H400FD pool/spa heater, Polaris PB4-60 booster pump, 280 pressure cleaner, Hayward Super II spa pump and Air Supply Silencer spa blower

  5. Back To Top    #5
    borjis's Avatar
    Join Date
    Aug 2014
    Location
    Pacific NW
    Posts
    2,049

    Re: Inconsistent FAS-DPD Test

    What test kit are you using?

    are you doing the 25 or 10 ml sample?

    How many scoops of powder?
    16x32 IG Vinyl, 13,000 Gals. Hayward S-244T sand filter, SP2810X15 pump
    (1.5 HP motor) Raypak PR266AEN (266k btu) gas heater. TF-100 Test Kit.
    Dolphin E10 Cleaning Robot.

  6. Back To Top    #6

    Join Date
    Jul 2015
    Location
    Clearwater, FL
    Posts
    83

    Re: Inconsistent FAS-DPD Test

    Taylor k2006 with reagents from tftestkits.com. 10 ml sample. 1 scoop of powder.
    16K gallon IG exposed aggregate pool with Aquarite SWG T15 cell, Hayward C1750 filter, EcoStar SP3400VSP pump, H400FD pool/spa heater, Polaris PB4-60 booster pump, 280 pressure cleaner, Hayward Super II spa pump and Air Supply Silencer spa blower

  7. Back To Top    #7

    Join Date
    Aug 2014
    Location
    Shallotte, NC
    Posts
    693

    Re: Inconsistent FAS-DPD Test

    Try 2 scoops of powder.... and the 25 ml sample....

  8. Back To Top    #8

    In the Industry

    duraleigh's Avatar
    Join Date
    Apr 2007
    Location
    Sebring, Florida
    Posts
    30,077

    Re: Inconsistent FAS-DPD Test

    What are you calling the endpoint of the test?
    Dave S.
    42k vinyl and concrete pool, 1.5hp pump, 140gpm filter
    TFTestkits , PoolMath , Pool School

  9. Back To Top    #9

    Join Date
    Jul 2015
    Location
    Clearwater, FL
    Posts
    83

    Re: Inconsistent FAS-DPD Test

    Endpoint is when the liquid is clear--when there is no tinge of red.
    16K gallon IG exposed aggregate pool with Aquarite SWG T15 cell, Hayward C1750 filter, EcoStar SP3400VSP pump, H400FD pool/spa heater, Polaris PB4-60 booster pump, 280 pressure cleaner, Hayward Super II spa pump and Air Supply Silencer spa blower

  10. Back To Top    #10

    Join Date
    Aug 2014
    Location
    Northern IL
    Posts
    4,615

    Re: Inconsistent FAS-DPD Test

    A meniscus determination can be problematic too. I sometimes use a small medical syringe to get an exact measurement of 10ml or 25ml when I start to doubt my consistency.

  11. Back To Top    #11
    aussieta's Avatar
    Join Date
    Jan 2014
    Location
    Langwarrin, Victoria, Australia
    Posts
    418

    Re: Inconsistent FAS-DPD Test

    if you take too long on the test it can turn pink again
    as mentioned one drop per second
    Extended Test Kit Directions
    http://www.taylortechnologies.com/Ch...P?ContentID=86
    SWG inground peanut shaped eco-bright concrete pool
    50,000 litres 13,000 gallons
    Lincoln Salt Chlorinator, 25 gram cell Sand Filter with glass media
    Taylor K2006-C test kit, Taylor K-1766 salt test, Dolphin M500 pool cleaner
    are you taking care of your pool or is the poolshop taking care of you

  12. Back To Top    #12

    In the Industry

    duraleigh's Avatar
    Join Date
    Apr 2007
    Location
    Sebring, Florida
    Posts
    30,077

    Re: Inconsistent FAS-DPD Test

    Endpoint is when the liquid is clear--when there is no tinge of red.
    That's pretty close. the endpoint is when there is no FURTHER color change to the sample. I.e. You add drops until the last drop you put in results in NO color change......subtract that drop and the result is the endpoint.

    That is the most common error for the FAS/DPD test but even then, you should not have a variation like you have. The test is predictable and precise and should consistently give you results within .5 ppm. Anything else indicates a testing error.
    Dave S.
    42k vinyl and concrete pool, 1.5hp pump, 140gpm filter
    TFTestkits , PoolMath , Pool School

  13. Back To Top    #13

    Join Date
    Jul 2015
    Location
    Clearwater, FL
    Posts
    83

    Re: Inconsistent FAS-DPD Test

    The reason I have repeated tests is to make sure I had not made errors in testing. I am confident that my testing process is good. That's why I am confused by the difference in results. I received new regents the other day, so I will see if they provide consistent results.
    16K gallon IG exposed aggregate pool with Aquarite SWG T15 cell, Hayward C1750 filter, EcoStar SP3400VSP pump, H400FD pool/spa heater, Polaris PB4-60 booster pump, 280 pressure cleaner, Hayward Super II spa pump and Air Supply Silencer spa blower

  14. Back To Top    #14

    In the Industry

    duraleigh's Avatar
    Join Date
    Apr 2007
    Location
    Sebring, Florida
    Posts
    30,077

    Re: Inconsistent FAS-DPD Test

    The reason I have repeated tests is to make sure I had not errored in testing.
    The chemistry on that test either works or doesn't work.....there is no "partially working" possibility I can think of. What do you think the issue may be?
    Dave S.
    42k vinyl and concrete pool, 1.5hp pump, 140gpm filter
    TFTestkits , PoolMath , Pool School

  15. Back To Top    #15

    Join Date
    Aug 2014
    Location
    Shallotte, NC
    Posts
    693

    Re: Inconsistent FAS-DPD Test

    Quote Originally Posted by duraleigh View Post
    The chemistry on that test either works or doesn't work.....there is no "partially working" possibility I can think of. What do you think the issue may be?
    I would suggest the use of insufficient R-0870 as the only "partially working" possibility.

    Also, there are the questions of:

    1) Taking the sample in the "wrong place," a place with poor circulation, too close to the top of the water, or too close to a skimmer.

    2) Taking a sample after the addition of chlorine, but before the water has mixed sufficiently.

    One major difficulty in offering help is that, not being able to see the whole process, we don't know whether there are one or more testing parameters affecting the result.

    Testing variance as large as 2.5 ppm, to me, suggests the likely possibility of multiple contributing factors....

  16. Back To Top    #16

    Join Date
    Jul 2015
    Location
    Clearwater, FL
    Posts
    83

    Re: Inconsistent FAS-DPD Test

    Quote Originally Posted by duraleigh View Post
    The chemistry on that test either works or doesn't work.....there is no "partially working" possibility I can think of. What do you think the issue may be?
    The only possibility I can imagine, then, is that the dropper on the bottle of R-0871 is creating inconsistently sized drops. I invert the bottle vertically upside down, release a drop, turn the bottle upright, then repeat. If I make any kind of mistake--release a double drop, let the drop contact the side of the cylinder so that all of it does not enter the sample, believe in any slight way I have lost count--I discard the test and start over.
    16K gallon IG exposed aggregate pool with Aquarite SWG T15 cell, Hayward C1750 filter, EcoStar SP3400VSP pump, H400FD pool/spa heater, Polaris PB4-60 booster pump, 280 pressure cleaner, Hayward Super II spa pump and Air Supply Silencer spa blower

  17. Back To Top    #17

    Join Date
    Jul 2015
    Location
    Clearwater, FL
    Posts
    83

    Re: Inconsistent FAS-DPD Test

    Quote Originally Posted by ewkearns View Post
    I would suggest the use of insufficient R-0870 as the only "partially working" possibility.

    Also, there are the questions of:

    1) Taking the sample in the "wrong place," a place with poor circulation, too close to the top of the water, or too close to a skimmer.

    2) Taking a sample after the addition of chlorine, but before the water has mixed sufficiently.

    One major difficulty in offering help is that, not being able to see the whole process, we don't know whether there are one or more testing parameters affecting the result.

    Testing variance as large as 2.5 ppm, to me, suggests the likely possibility of multiple contributing factors....
    Sample collections is sound: inverted sample bottle into the deep end of the pool up to my elbow, turn the bottle right side up and fill, cover top of bottle with palm and bring it to the surface. Pump has been running for hours, since I use an SWG, I stay away from returns when I take the sample, and I am nowhere near the skimmer. Besides, issues with sample collection should not affect testing inconsistency from the same sample.
    16K gallon IG exposed aggregate pool with Aquarite SWG T15 cell, Hayward C1750 filter, EcoStar SP3400VSP pump, H400FD pool/spa heater, Polaris PB4-60 booster pump, 280 pressure cleaner, Hayward Super II spa pump and Air Supply Silencer spa blower

  18. Back To Top    #18

    Join Date
    Aug 2014
    Location
    Shallotte, NC
    Posts
    693

    Re: Inconsistent FAS-DPD Test

    Have you tried this with the 25 mL sample size, using 2 scoops of R-0870?

  19. Back To Top    #19
    JoyfulNoise's Avatar
    Join Date
    May 2015
    Location
    Tucson, AZ
    Posts
    11,579

    Re: Inconsistent FAS-DPD Test

    Quote Originally Posted by ThomasG View Post
    The only possibility I can imagine, then, is that the dropper on the bottle of R-0871 is creating inconsistently sized drops. I invert the bottle vertically upside down, release a drop, turn the bottle upright, then repeat. If I make any kind of mistake--release a double drop, let the drop contact the side of the cylinder so that all of it does not enter the sample, believe in any slight way I have lost count--I discard the test and start over.
    I would suggest you leave the bottle inverted over the sample tube and just keep the drops dripping out. You should not be inverting & righting the dropper bottle during the testing. If you watch all of the videos on the Taylor website (see the 4:30 mark on this video), you'll notice the fellow that does the tests is very skilled - he can swirl with one hand while simultaneously adding drops of reagent with the other hand. That's what you need to achieve. If you can't get to that level, then you should purchase a SpeedStir as that will free you from having to swirl the test solution.

    I also echo ewkearns. Too often folks try to "save" powder by only using one scoop. While I understand the reagents are expensive and they take time to ship, using less than the required amount of powder can lead to measurement error. If you read the Taylor instructions, they clearly specify that you "add two dippers of powder to the test solution". That is two (2) dippers, FULL STOP. You do not modify the amount of powder regardless of the sample size. The R-0870 powder is actually a mixture of several different chemicals, not just the DPD dye. The chemical makeup is designed to adjust the pH of the test sample to right range and buffer the pH during the test. There are also chemicals in there to protect against interferences from metals. So using less than the prescribed amount of R-0870 can possibly lead to trouble. There's a lot of anecdotal stories on TFP about being able to use less powder to "save" on reagents; I would assert that you should simply follow the instructions as-written.

    In fact, I actually use the Taylor Unit-Dose Dispenser for my R-0870 powder which dispenses a little bit more than two full dippers worth of powder. Do I use more powder than average, sure. Do I ever get inconsistent test results ... not yet.
    Matt
    16k IG PebbleTec pool, 650gal spa, spillway and waterfall, 3HP IntelliFlo VS / 1.5HP WhisperFlo, Pentair QuadDE-100 filter, IC40 SWCG, MasterTemp 400k BTU/hr NG heater, KreepyKrauly suction-side cleaner Dolphin S300i robot, EasyTouch controls, city water, K-1001, K-2006 and K-1766 test kits, Mannitol test for borates

  20. Back To Top    #20

    Join Date
    Jul 2015
    Location
    Clearwater, FL
    Posts
    83

    Re: Inconsistent FAS-DPD Test

    Quote Originally Posted by JoyfulNoise View Post
    I would suggest you leave the bottle inverted over the sample tube and just keep the drops dripping out. You should not be inverting & righting the dropper bottle during the testing. If you watch all of the videos on the Taylor website, you'll notice the fellow that does the tests is very skilled - he can swirl with one hand while simultaneously adding drops of reagent with the other hand. That's what you need to achieve. If you can't get to that level, then you should purchase a SpeedStir as that will free you from having to swirl the test solution.

    I also echo ewkearns. Too often folks try to "save" powder by only using one scoop. While I understand the reagents are expensive and they take time to ship, using less than the required amount of powder can lead to measurement error. If you read the Taylor instructions, they clearly specify that you "add two dippers of powder to the test solution". That is two (2) dippers, FULL STOP. You do not modify the amount of powder regardless of the sample size. The R-0870 powder is actually a mixture of several different chemicals, not just the DPD dye. The chemical makeup is designed to adjust the pH of the test sample to right range and buffer the pH during the test. There are also chemicals in there to protect against interferences from metals. So using less than the prescribed amount of R-0870 can possibly lead to trouble. There's a lot of anecdotal stories on TFP about being able to use less powder to "save" on reagents; I would assert that you should simply follow the instructions as-written.

    In fact, I actually use the Taylor Unit-Dose Dispenser for my R-0870 powder which dispenses a little bit more than two full dippers worth of powder. Do I use more powder than average, sure. Do I ever get inconsistent test results ... not yet.
    I'll try that.